[time-nuts] Am I doing something wrong?

Bill Beam wbeam at gci.net
Thu Sep 26 20:00:02 UTC 2019


It looks like since last Oct the osc jumped frequency and has become less stable.
This is not unheard of for an ocxo.  It may be time to replace with a true
Trimble Thunderbolt.  A couple are now on ebay for 2-300$.  It is possible to
replace the ocxo, assuming nothing else is wrong with the circuit.

On Thu, 26 Sep 2019 13:21:25 -0400, Roger Rehr W3SZ wrote:

>Hi All,

>Sorry I forgot to attach the freq vs DAC voltage table to the email that 
>I just sent:
>DAC V-  Freq
>+5.0- - - -  9.999.961
>+0.0- - - -  9.999.985
>-3.0- - - -  9.999.998
>-3.5- - - -  10.000.000
>-3.8- - - -  10.000.001
>-4.0- - - -  10.000.002
>-4.5- - - -  10.000.004
>-5.0- - - -  10.000.006

>So the oscillator is not "centered" on 10.000.000 anymore.

>73,
>Roger
>W3SZ

>On 9/26/2019 1:06 PM, Roger Rehr W3SZ wrote:
>> Hi Nigel, and Bob, and Bill!
>>
>> Thanks for the notes!-  I very much appreciate your helpful comments.-  
>> I will give my responses to all of your helpful posts here.-  Sorry for 
>> the delay in responses.-  The reason for the delay is that my 
>> Thunderbolt is at a remote site and I can communicate with it via Lady 
>> Heather remotely, but I had to drive out to the site to do the tests 
>> with an HP 5342A frequency counter and pulse analysis of the pps 
>> port.-  I drove out to the site today.
>>
>> First, Bob and Bill,
>>
>> 1.
>> DAC set to +0.5 V gives significant frequency errors:-  HP counter 
>> reads 9.999.984 MHz and osc error as reported by LH is 1,68x,xxx.xx 
>> ppt.-  Interestingly, at this voltage pps error is flipping between +50 
>> and -50.
>>
>> The DAC voltage that minimizes both counter error and pps/osc error as 
>> reported by LH is around -3.553 V.-  This is dramatically different 
>> than the setting last October when all was working well, as you know.
>>
>> With DAC voltage -3.553 V the osc error and pps error are reduced, but 
>> they jump around quite a bit:
>> pps short-term variation is for example between about +50 and -50 and 
>> osc short term variation is between about +5000 and -5000.-  HP counter 
>> reads 10.000.000 MHz.
>>
>> Even when I have the DAC close enough to optimum that my counter reads 
>> 10.000.000 MHz and the osc error is flipping around both sides of 
>> zero, the unit stays in "Power-up" mode and the DAC voltage stays at 
>> wherever I set it (either using d>>s or &>>i and then power cycle) and 
>> does not vary.
>>
>>
>> 2.
>> I get a 10 usec duration +5V pulse every 1.00 seconds from the PPS 
>> port, as per specifications.-  I cannot give you greater precision than 
>> that on the pulse interval.
>>
>>
>> Nigel,
>>
>> I agree that something appears odd with this unit that had been 
>> running with no problems for at least a year at my location.-  There 
>> have been no known system / config changes by me, although there is 
>> always the chance that I fat-fingered something when playing with the 
>> Lady Heather display at some point.
>>
>> The receiver self identifies as a TSIP (Trimble TSIP Binary) receiver 
>> when Lady Heather starts up.-  The receiver type is not specified in my 
>> cfg file, which instead has the default line /rx which as you know 
>> means auto-detect receiver type.
>>
>> I had assumed that the "Nortel NTGS/NTBW" designation that you mention 
>> and that appears on the Lady Heather parameter display had something 
>> to do with Trimble having supplied the unit I now have to Nortel and 
>> this "Nortel NTGS/NTBW" designation being a custom ID placed in the 
>> firmware so that the unit would be "branded" as Nortel.-  This 
>> supposition is consistent with line 2170 of the heather.txt file which 
>> says "Trimble telecom GPSDOs (NTBW,NTPX,NTGS,etc)".-  There is also a 
>> line in the heather.pdf file that says, "Note that non-Thunderbolt 
>> GPSDOs (such as the NTBW and NTPx telecom devices) do not all support 
>> saving disciplining parameters to EEPROM and some do not allow users 
>> to change the disciplining parameters. You can use "/d" command line 
>> options for setting the PLL parameters from the command line. If you 
>> run the auto-tune command on non-Thunderbolt devices then only the 
>> satellite elevation mask (and perhaps the signal level mask) values 
>> will be set".
>>
>> The physical unit is labeled "Trimble THUNDERBOLT", so I guess, 
>> looking at all of the data, that I have a Trimble / Nortel device.
>>
>> But the issue of whether it is a Nortel/Trimble Thunderbolt or just a 
>> Thunderbolt wouldn't cause the current problem that I am having now in 
>> a unit that previously worked perfectly for me.
>>
>> Just so you know that my heather.cfg file is "OK" and not responsible 
>> for any problems, here it is:
>>
>> /9
>> /rx
>> /br=9600:8:n:1
>> /ro=1*
>> /tz=-5EST/EDT
>> /b=1
>> /tb=W3SZ
>> /gm
>> /gw=0
>> /gqw
>> /gcs-
>> /gcg
>>
>> To further document that this issue is not a "setup" problem, I did 
>> another factory reset of the unit and then did not use Lady Heather 
>> but only monitored function with TBoltMon.exe and I did not do any 
>> setup with that utility until the self survey was done and more than 
>> two hours had passed, and then I only set and saved the position. 
>> Function and-  results were unchanged, and the unit was still stuck in 
>> Power-Up mode and DAC voltage was still stuck at zero, but everything 
>> else was normal according to the TBoltMon readouts.
>>
>> Thanks again and 73,
>>
>> Roger
>> W3SZ
>>


>_______________________________________________
>time-nuts mailing list -- time-nuts at lists.febo.com
>To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
>and follow the instructions there.