Hi Bob
yes sure I know. At least my homebrew GPSDO and the STAR4 GPSDO I have are
always powered, but since they don't take much space on my desk that's
okay. But for the signal generators (or SpecAn etc) it is a completely
different story; I usually take them from the shelf when I need them and
put them back afterwards, because they simply take too much space away. And
my home lab has only limited space, unfortunately. I am thinking since
months about a better setup (how arrange everything, how to stack my
equipment and such) but it is not so easy because simply stacking all
equipment on top of each other is unsatisfactory (in that case one needs in
general the instrument at the very top, which is simply too far away :-)).
The 8663A is a nice machine, but it would fill my desk already by 50%, so
no way to leave it there (and therefore I cannot keep it powered).
Tobias
On Tue, Apr 14, 2020 at 1:54 PM Bob kb8tq <kb8tq at n1k.org> wrote:
> Hi
>
> > On Apr 14, 2020, at 2:31 AM, Tobias Pluess <tpluess at ieee.org> wrote:
> >
> > Hi Taka
> > good to know you're working on the same stuff :-)
> > I also think that this is the goal, sort of. But apparently there are
> some
> > pitfalls. I hope the experts on this type of measurement (Bob) can give
> > more hints ;-)
> >
> > Btw good to know your 10811 took 2 months to settle. In that case I will
> > try to give mine also more time from power up until I do measurements.
> But
> > keeping the 8663 powered on is almost no option for me - my homelab is
> > simply too small and from time to time I therefore have to tidy it up a
> bit
> > and put some of the equipment in another shelf.
>
> You will need the ability to keep stuff on power. That’s true of pretty
> much
> any frequency standard. Indeed the compromise is often made with a Cs
> standard since they only have just so many running hours before they
> run out of “fuel”. Anything that’s OCXO or Rb based …. leave it on power.
>
> One of the side effects of being locked down is that there is *lots* of
> time
> to pick up the clutter and *really* get organized. It’s a amazing just how
> many weeks that sort of work has soaked up here :)
>
> Bob
>
> >
> >
> > Tobias
> >
> >
> > On Tue., 14 Apr. 2020, 01:19 Taka Kamiya via time-nuts, <
> > time-nuts at lists.febo.com> wrote:
> >
> >> I am working on pretty much the same thing. My HP105B has an HP11801
> >> inside. There is no telling how long it has been off. Frequency
> wandered
> >> ALL OVER THE PLACE. It took 2 months of continuous operation to settle
> >> down and just do the normal aging/drifting.
> >>
> >> I have a same question as you do. How come not just one HP5335? Your
> >> test setup has two output. One goes to start, the other goes to stop.
> >> Measure the time interval. Isn't that the goal?
> >>
> >> ---------------------------------------
> >> (Mr.) Taka Kamiya
> >> KB4EMF / ex JF2DKG
> >>
> >>
> >> On Monday, April 13, 2020, 5:53:52 PM EDT, Bob kb8tq <kb8tq at n1k.org>
> >> wrote:
> >>
> >> Hi
> >>
> >>> On Apr 13, 2020, at 5:06 PM, Tobias Pluess <tpluess at ieee.org> wrote:
> >>>
> >>> Hi Bob
> >>> awesome, thanks! of course it is 1e6, not 1e7, I got a mistake :-)
> >>>
> >>> Maybe I have some good OpAmps for this purpose in my box. I will try
> it!
> >> of
> >>
> >> You need something that is quiet (like the OP-37) and has a pretty good
> >> slew
> >> rate. Past that, there are a lot of candidates. The TI OPA-228 family is
> >> one.
> >>
> >>> course I saw that my setup was not ideal as there was a bit of noise on
> >> the
> >>> signals which I guess does lead to some jitter in the trigger circuit
> and
> >>> therefore decreases my measurement noise floor.
> >>
> >> Typically a good limiter takes you from 3 or 4 digits up to 6 or 7 good
> >> digits.
> >> Net result is a measurement that’s good in the vicinity of parts in
> 10^-13
> >>
> >>>
> >>> Can you say something about how it would be done using a TIC?
> >>> I don't have two identically good counters, but the HP 5335A could be
> >> used
> >>> as TIC, couldn't it.
> >>
> >> The standard way of doing the test is to run two counters / two TIC/s /
> >> two whatever’s.
> >> I know of no practical way to do it with a single 5335.
> >>
> >>>
> >>> And the offset source I used is not directly the HP 10811, but the HP
> >> 8663A
> >>> Signal generator internally uses a 10811 as reference source. But I
> >> didn't
> >>> wait for days for it to warm up properly. (Should I?)
> >>
> >> The 8663 synthesizer adds a *lot* of crud to the 10811. Regardless of
> how
> >> you
> >> use the 10811, it needs to be on for a while. How long very much depends
> >> on
> >> just how long it’s been off. Best to keep it on all the time.
> >>
> >>>
> >>>> Fun !!!
> >>> Yea, of course! :-)
> >>> I already implemented the ADEV, MDEV and TDEV calculations in Matlab by
> >>> myself. I use TimeLab to see what numbers I should expect, and then I
> >> want
> >>> to compute it all myself in Matlab because I want to see how it
> actually
> >>> works. ;-)
> >>
> >> Be careful any time you code this stuff for the first time. It’s
> amazingly
> >> easy
> >> ( = I’ve done it ….) to make minor errors. That’s in no way to suggest
> that
> >> you should not code it up yourself. I generally do it in Excel or in C.
> >>
> >> Bob
> >>
> >>>
> >>>
> >>> Best
> >>> Tobias
> >>>
> >>>
> >>>
> >>> On Mon, Apr 13, 2020 at 10:50 PM Bob kb8tq <kb8tq at n1k.org> wrote:
> >>>
> >>>> Hi
> >>>>
> >>>> Ok, first the math:
> >>>>
> >>>> If your offset oscillator is 10 Hz high at 10 MHz, you have a:
> >>>>
> >>>> 10,000,000 / 10 = 1,000,000 : 1 multiplier in front of the DMTD
> >>>>
> >>>> You get to add a 6 to what Time Lab shows you.
> >>>>
> >>>> If you are getting an ADEV at 1 second of 1x10^-4 then that multiplier
> >>>> gets you to 1x10^-10
> >>>>
> >>>> So, what’s going on?
> >>>>
> >>>> You can’t feed the mixer outputs straight into a counter. The counter
> >>>> front
> >>>> end does not handle LF audio sine waves very well. You need to do an
> >>>> op-amp based limiter. A pair of OP-37’s in each leg ( or something
> >>>> similar)
> >>>> should do the trick.
> >>>>
> >>>> Second, the offset source needs to be pretty good. A 10811 tuned high
> >> with
> >>>> both the mechanical trim and the EFC is a pretty good choice to start
> >> out.
> >>>>
> >>>> If you only have one counter, simply ignore the second channel. You
> are
> >> now
> >>>> running a single mixer. It still works as a comparison between the
> >> offset
> >>>> oscillator
> >>>> and your DUT.
> >>>>
> >>>> If you want to do it properly as a DMTD, then you set up two counters.
> >> One
> >>>> to measure mixer A and the other to measure mixer B. Set them both up
> >> to
> >>>> measure frequency. Time tag the data files so you know which reading
> >>>> matches up with which.
> >>>>
> >>>> Fun !!!
> >>>>
> >>>> Bob
> >>>>
> >>>>> On Apr 13, 2020, at 3:18 PM, Tobias Pluess <tpluess at ieee.org> wrote:
> >>>>>
> >>>>> Hi again Bob
> >>>>>
> >>>>> I tried to do some measurements with a DMTD!
> >>>>> In my junk box I found a little PCB from earlier experiments on that
> >>>> topic,
> >>>>> with a power splitter and two SRA-3H mixers, it was even already
> wired
> >>>> for
> >>>>> the DMTD configuration. So I gave it a try!
> >>>>> As "transfer oscillator" I used my HP 8663A signal generator, and set
> >> it
> >>>>> high in frequency by 10 Hz. To the two mixers, I connected the two
> >> 10MHz
> >>>>> signals and at the mixer outputs, I put a little lowpass filter with
> >>>> 100Hz
> >>>>> corner frequency.
> >>>>> The output signals from the two SRA-3 mixers are almost 0.5Vpp, so I
> >>>> tried
> >>>>> to feed them directly into the HP 5335A TIC and used the TI mode to
> >>>> measure
> >>>>> the delay between the two signals.
> >>>>> This gives 10 readings/sec, which I try to process with TimeLab.
> >>>>> It does give some interesting graphs, but I don't know yet how to
> >>>> correctly
> >>>>> set up TimeLab for this kind of measurement. I.e. now, I get an ADEV
> in
> >>>> the
> >>>>> order of 1e-4 (at tau=1sec) to 1e-5 (at tau=500sec). So does that
> mean
> >> I
> >>>>> simply need to multiply this with 1e-7 to get the *real* ADEV at
> 10MHz?
> >>>>> this would mean that my real ADEV is in the range of 1e-11 to 1e-12,
> >>>> which
> >>>>> is indeed my target value, BUT I expect that things are not that
> >> simple.
> >>>>> (i.e. what if I didn't set the transfer oscillator high by +10Hz but
> >> only
> >>>>> by 9.9Hz for example).
> >>>>> Can you give some hints on that?
> >>>>> Of course I also did the noise floor test (i.e. I fed the 10MHz
> signal
> >>>> into
> >>>>> a power splitter and connected the two outputs to my DMTD with two
> >>>>> different lenghts of cables. This gave results starting at 1e-4 going
> >>>> down
> >>>>> to 1e-7, maybe it would have gone even lower but I measured only for
> a
> >>>>> couple of minutes.)
> >>>>>
> >>>>> Can you give some hints on that?
> >>>>>
> >>>>> Best
> >>>>> Tobias
> >>>>> HB9FSX
> >>>>>
> >>>>>
> >>>>>
> >>>>>
> >>>>> On Fri, Apr 3, 2020 at 1:45 PM Bob kb8tq <kb8tq at n1k.org> wrote:
> >>>>>>
> >>>>>>> Hi
> >>>>>>>
> >>>>>>> The quick way to do this is with a single mixer. Take something
> like
> >> an
> >>>>>>> old
> >>>>>>> 10811 and use the coarse tune to set it high in frequency by 5 to
> 10
> >>>> Hz.
> >>>>>>>
> >>>>>>> Then feed it into an RPD-1 mixer and pull out the 5 to 10 Hz audio
> >>>> tone.
> >>>>>>> That tone is the *difference* between the 10811 and your device
> under
> >>>>>>> test.
> >>>>>>> If the DUT moves 1 Hz, the audio tone changes by 1 Hz.
> >>>>>>>
> >>>>>>> If you measured the 10 MHz on the DUT, that 1 Hz would be a very
> >> small
> >>>>>>> shift
> >>>>>>> ( 0.1 ppm ). At 10 Hz it’s a 10% change. You have “amplified” the
> >>>> change
> >>>>>>> in frequency by the ratio of 10 MHz to 10 Hz ( so a million X
> >> increase
> >>>> ).
> >>>>>>>
> >>>>>>> *IF* you could tack that on to the ADEV plot of your 5335 ( no,
> it’s
> >>>> not
> >>>>>>> that
> >>>>>>> simple) your 7x10^-10 at 1 second would become more 7x10^-16 at 1
> >>>>>>> second.
> >>>>>>>
> >>>>>>> The reason its not quite that simple is that the input circuit on
> the
> >>>>>>> counter
> >>>>>>> really does not handle a 10 Hz audio tone as well as it handles a
> 10
> >>>> MHz
> >>>>>>> RF signal. Instead of getting 9 digits a second, you probably will
> >> get
> >>>>>>> three
> >>>>>>> *good* digits a second and another 6 digits of noise.
> >>>>>>>
> >>>>>>> The good news is that an op amp used as a preamp ( to get you up to
> >>>> maybe
> >>>>>>> 32 V p-p rather than a volt or so) and another op amp or three as
> >>>>>>> limiters will
> >>>>>>> get you up around 6 or 7 good digits. Toss in a cap or two as a
> high
> >>>> pass
> >>>>>>> and low pass filter ( DC offsets can be a problem ….) and you have
> a
> >>>>>>> working
> >>>>>>> device that gets into the parts in 10^-13 with your 5335.
> >>>>>>>
> >>>>>>> It all can be done with point to point wiring. No need for a PCB
> >>>> layout.
> >>>>>>> Be
> >>>>>>> careful that the +/- 18V supplies to the op amp *both* go on and
> off
> >> at
> >>>>>>> the
> >>>>>>> same time ….
> >>>>>>>
> >>>>>>> Bob
> >>>>>>>
> >>>>>>
> >>>>> _______________________________________________
> >>>>> time-nuts mailing list -- time-nuts at lists.febo.com
> >>>>> To unsubscribe, go to
> >>>> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
> >>>>> and follow the instructions there.
> >>>>
> >>>>
> >>>> _______________________________________________
> >>>> time-nuts mailing list -- time-nuts at lists.febo.com
> >>>> To unsubscribe, go to
> >>>> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
> >>>> and follow the instructions there.
> >>>>
> >>> _______________________________________________
> >>> time-nuts mailing list -- time-nuts at lists.febo.com
> >>> To unsubscribe, go to
> >> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
> >>> and follow the instructions there.
> >>
> >>
> >> _______________________________________________
> >> time-nuts mailing list -- time-nuts at lists.febo.com
> >> To unsubscribe, go to
> >> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
> >> and follow the instructions there.
> >>
> >> _______________________________________________
> >> time-nuts mailing list -- time-nuts at lists.febo.com
> >> To unsubscribe, go to
> >> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
> >> and follow the instructions there.
> >>
> > _______________________________________________
> > time-nuts mailing list -- time-nuts at lists.febo.com
> > To unsubscribe, go to
> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
> > and follow the instructions there.
>
>
> _______________________________________________
> time-nuts mailing list -- time-nuts at lists.febo.com
> To unsubscribe, go to
> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
> and follow the instructions there.
>
More information about the Time-nuts_lists.febo.com
mailing list