[time-nuts] Time-Nutters-- Adding Rubidium to a Thunderbolt...?
Poul-Henning Kamp
phk at phk.freebsd.dk
Mon Mar 2 08:42:36 UTC 2020
--------
In message <CADHrwpfp1kuXWW8XzQpqK+m=47Z5PN4suMzfve_GjcjhZzDRvA at mail.gmail.com>, Dana Whitlow writes:
>Probably the easiest solution would be to feed 1 PPS from the T'bolt into a full-featured
>PRS-10 Rb, then fiddle with the loop parameters in the PRS-10 to best suit your needs.
Been there, tried that, not a good idea.
The PPS input of the PRS-10 is OK for "coarse" PPS signals like
those you get out of a GPS, because the sawtooth noise dithers over
the unlinearity and "missing codes" of the PRS-10 timestamping
circuit.
When you feed a PRS-10 a "perfect" PPS signal, you get really
erratic behaviour.
Only way to slave a PRS-10 from a "good" source, is to do
phase-comparison on the 10MHz signal, and after suitable
filtering and integration, use that to set the PRS-10 frequency
via the serial port.
That, on the other hand can give amazing results because of the
relatively good X-tals they put in the PRS-10.
--
Poul-Henning Kamp | UNIX since Zilog Zeus 3.20
phk at FreeBSD.ORG | TCP/IP since RFC 956
FreeBSD committer | BSD since 4.3-tahoe
Never attribute to malice what can adequately be explained by incompetence.