[time-nuts] HP5065a cfield issue

paul swed paulswedb at gmail.com
Thu Dec 3 16:47:47 UTC 2020


Just rechecked everything as the system seems far to good to be true. Yet
everything is actually correct. It is drifting fast compared to the TBolt.
But its very small. I am still in the 1.5-1.7ns range after 2 hours and 43
minutes. I believe by calculation that 1.5 E-13. That makes no sense.
Simply taking 1.5ns/9,780 seconds.
Regards
Paul.

On Thu, Dec 3, 2020 at 11:15 AM paul swed <paulswedb at gmail.com> wrote:

> Hi Poul
> At least the front panel meter readings do not show any changes. Perhaps
> an increase of 1 count sub tick mark changes that I would attribute to
> typical drift. Looking over years the readings are consistent. Will say the
> second harmonic level is high at 55. But its been that high in the past and
> at other times as low as 28. But on average 48-50.
> Regards
> Paul
>
> On Thu, Dec 3, 2020 at 10:51 AM Poul-Henning Kamp <phk at phk.freebsd.dk>
> wrote:
>
>> --------
>>
>> Could it be the thermostat circuits have shifted so you run at a
>> different temperature now ?
>>
>> --
>> Poul-Henning Kamp       | UNIX since Zilog Zeus 3.20
>> phk at FreeBSD.ORG         | TCP/IP since RFC 956
>> FreeBSD committer       | BSD since 4.3-tahoe
>> Never attribute to malice what can adequately be explained by
>> incompetence.
>>
>