Hi Dave,

I replaced the battery, went through the procedure but without the
extender card. Turned out that the interpolator calibration was good
enough, so it didn't take much time and it worked great since (well,
except for when the CRT failed, but that was fairly trivial fix in itself).

Just follow the instructions in the service manual, not too hard. We
have not had a reason to distrust the measurements made.

A somewhat quirky instrument that does some stuff like few other
stand-alone instruments, and gets regular use in the lab at work, in
addition to the 53132A counter + TimeLab setup I run.

Good luck!

Cheers,
Magnus

On 2020-06-14 00:52, Dave Miller wrote:
> My unit arrived this week and it has as advertised the  Error 160  - Out of
> sensisty Cal. error.
> According to the manual you have to go through a whole series of
> adjustments after you change out the battery.
> It also says you have to use a 05372-60016 exetender card. Can anyone
> provide me details on what this is? I can probably make one if I had a
> picture or some details.
>
> Also do you really have to run all the tests or just the first one in
> section 3-30?
>
> Even with this error it seems to make measurements that are believable.
>
> Thanks
> Dave
> VE7HR
>
>
>
>
>