[time-nuts] Re: What phase variations to expect in a DMTD due to temperature fluctuations?
Erik Kaashoek
erik at kaashoek.com
Fri Oct 28 12:49:23 UTC 2022
Hi John,
Thanks for the practical advice and the measurements.
After ironing out some problems the measurement noise with two equal
inputs and a tau of 0.1 s is now below 100 fs.
Long term phase stability has also improved and typical slow phase
variations over an hour, when measured on a desktop in a room with
people, are now within 4 ps
The next step may be to replace the current plastic enclosure with a
decommissioned fridge, but that would defeat the idea of having a small
battery powered portable device.
Erik.
On 24-10-2022 21:49, John Miles via time-nuts wrote:
> You should see similar sub-picosecond drift figures from your own
> DMTD, given similar conditions. If not, there will be a reason, one that is
> probably not too hard to address.
>
>