[volt-nuts] Keithley 2182A
Russ Ramirez
russ.ramirez at gmail.com
Sun Jun 26 17:03:01 EDT 2016
Thanks. I did not realize that the CPLD was for ADC control until I re-read
Illya's article again, and I then realized with the parts complement
present that the "ADC" must be an integrating type. Next time I crack it
open, I'll have more to check.
Russ
On Sun, Jun 26, 2016 at 3:44 PM, Todd Micallef <tmicallef at gmail.com> wrote:
> The repair was performed last November and my memory is bad, too many
> projects since then. The two defective parts were the TF-245 and U626.
>
> My symptoms were noisy readings across all ranges. I could not get them to
> settle down even with the input shorted. I measured the TF-245 in-circuit
> to see if anything was too far out to be in spec. I was certain that the
> 30K resistor between 5-11 was open, so I removed it and measured. There was
> also another open 90K resistor between 7-8.
>
> I also had to trace an input signal through the meter and it led to the
> resistor array. I will try and find the actual in-circuit measurements
> before the part was replaced.
>
> I
>
>
> Todd
>
> On Sun, Jun 26, 2016 at 4:23 PM, Russ Ramirez <russ.ramirez at gmail.com>
> wrote:
>
> > Looking forward to your notes Todd. A bit more info on the repair.
> >
> > DCV1 is OK except for the 10 V range. A test voltage input of 100 mV
> > appears on the input side of Q665 fine, but not on the output side in the
> > 10 V range, however a DMM measurement is solid in the other applicable
> > ranges. So this issue does seem to be completely isolated to the 10 V
> > range.
> >
> > DCV2 input is fine through the path to the ADC (TP652) in all ranges, but
> > the unit is not making a stable measurement. For example, a 9 V input
> reads
> > between 8.8 V and 9.2 V on the 10 V scale. The same 9 V input to DCV1
> also
> > looks good up to TP652, on the 100 V range since 10 v is out it, and it
> > reads a stable 9.002 V at that TP. Without a schematic, it will take some
> > time to determine what is different after the TP652 point - maybe there
> are
> > in fact 2 ADCs, but it's not obvious either way.
> >
> > Russ
> >
> > On Sun, Jun 26, 2016 at 11:16 AM, Todd Micallef <tmicallef at gmail.com>
> > wrote:
> >
> > > I have to review my notes, but I know of the two bad parts that were
> > > found, one was the TF-245. I happened to have a spare on hand. I will
> > look
> > > through any notes to get the other part number.
> > >
> > > Todd
> > >
> > > Sent from my iPad
> > >
> > > > On Jun 26, 2016, at 10:09, Russ Ramirez <russ.ramirez at gmail.com>
> > wrote:
> > > >
> > > > Thanks Illya - very nice piece of work!
> > > >
> > > > Russ
> > > >
> > > >> On Sun, Jun 26, 2016 at 5:27 AM, Illya Tsemenko <illya at xdevs.com>
> > > wrote:
> > > >>
> > > >> If you looking into repair or internals detail of 2182, this draft
> > > article
> > > >> may help: https://xdevs.com/fix/kei2182/
> > > >> Keithley stopped releasing schematics and detailed information since
> > > >> xxxx-series instruments due to IP issues.
> > > >>
> > > >> 26/06/16 18:08, Brian Smith пишет:
> > > >>
> > > >> Hello Russ
> > > >>>
> > > >>> It appears that there are neither a service manual nor a schematic
> > > >>> available anywhere.
> > > >>>
> > > >>> My educated guess is that this is because the 2182A is a current
> > model
> > > >>> instrument. The newest production model keithley gear does not have
> > > >>> associated service manuals nor schematics published. These appear
> to
> > be
> > > >>> treated as trade secrets.
> > > >>>
> > > >>> You can have your 2182A serviced by Keithley if it is in need of
> > > repair.
> > > >>>
> > > >>> I had my model 2000 DMM that I bought from eBay serviced by
> Keihtley
> > > for
> > > >>> $560. They replaced the motherboard bringing the DMM up to brand
> new
> > > >>> condition. I asked for and received the calibration data with the
> > > returned
> > > >>> instrument.
> > > >>>
> > > >>> best regards,
> > > >>>
> > > >>> Brian
> > > >>>
> > > >>>
> > > >>>
> > > >>> On Fri, Jun 24, 2016 at 10:38 PM, Russ Ramirez <
> > russ.ramirez at gmail.com
> > > >
> > > >>> wrote:
> > > >>>
> > > >>> The service manual, or a schematic. I asked just in case someone
> had
> > > one
> > > >>>> from an original purchase. I also asked a friend who worked for
> > > Keithley,
> > > >>>> but apparently no one he knew had one either. Hard to find.
> > > >>>>
> > > >>>> Russ
> > > >>>> On Jun 24, 2016 6:26 PM, "Brian Smith" <briansmith23456 at gmail.com
> >
> > > >>>> wrote:
> > > >>>>
> > > >>>> What document do you need? Keithley publishes most needed docs.
> > > >>>>> On Jun 24, 2016 12:59 PM, "Russ Ramirez" <russ.ramirez at gmail.com
> >
> > > >>>>> wrote:
> > > >>>>>
> > > >>>>> Just wondering if any of you might have a PDF of the subject
> > > >>>>>>
> > > >>>>> Nanovoltmeter?
> > > >>>>>
> > > >>>>>> TIA
> > > >>>>>>
> > > >>>>>> Russ
> > > >>>>>> _______________________________________________
> > > >>>>>> volt-nuts mailing list -- volt-nuts at febo.com
> > > >>>>>> To unsubscribe, go to
> > > >>>>>> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > > >>>>>> and follow the instructions there.
> > > >>>>>>
> > > >>>>>> _______________________________________________
> > > >>>>> volt-nuts mailing list -- volt-nuts at febo.com
> > > >>>>> To unsubscribe, go to
> > > >>>>> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > > >>>>> and follow the instructions there.
> > > >>>>>
> > > >>>>> _______________________________________________
> > > >>>> volt-nuts mailing list -- volt-nuts at febo.com
> > > >>>> To unsubscribe, go to
> > > >>>> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > > >>>> and follow the instructions there.
> > > >>>>
> > > >>>> _______________________________________________
> > > >>> volt-nuts mailing list -- volt-nuts at febo.com
> > > >>> To unsubscribe, go to
> > > >>> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > > >>> and follow the instructions there.
> > > >>>
> > > >>
> > > >>
> > > >>
> > > >> _______________________________________________
> > > >> volt-nuts mailing list -- volt-nuts at febo.com
> > > >> To unsubscribe, go to
> > > >> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > > >> and follow the instructions there.
> > > >>
> > > > _______________________________________________
> > > > volt-nuts mailing list -- volt-nuts at febo.com
> > > > To unsubscribe, go to
> > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > > > and follow the instructions there.
> > > _______________________________________________
> > > volt-nuts mailing list -- volt-nuts at febo.com
> > > To unsubscribe, go to
> > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > > and follow the instructions there.
> > >
> > _______________________________________________
> > volt-nuts mailing list -- volt-nuts at febo.com
> > To unsubscribe, go to
> > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts