[time-nuts] Morion MV89A position
Magnus Danielson
magnus at rubidium.dyndns.org
Sat Jan 25 13:24:12 UTC 2014
On 24/01/14 09:12, Azelio Boriani wrote:
> This issue has been discussed at length several times. Too much
> insulation is not necessarily desirable since the oscillator power has
> to be dissipated and the oven design of the MV89 wasn't meant to be
> put in a dewar. Better reduce the temperature gradient then fully
> isolate the OCXO. The upside-down position is not an issue in my
> opinion.
Running it upside-down is by itself not an issue, but do expect that the
frequency will shift when you turn it over and do expect that it will be
starting a new drift period from the new frequency until it has relaxed.
Always be careful with isolation and thermal loading, as the oven loop
can turn nasty on you if you are not careful.
Cheers,
Magnus