[time-nuts] PRS10
Mike Cook
michael.cook at sfr.fr
Thu Aug 11 16:07:09 UTC 2016
> Le 11 août 2016 à 12:15, Martyn Smith <martyn at ptsyst.com> a écrit :
>
> Hello,
>
> I have a strange fault on a PRS10. Wondered if anyone had seen it.
>
> I am running a PRS10 and at the same time monitoring its FC (the frequency control value) and Status byte data values.
>
> So normally, the Status shows are all zeros, except 2nd from last is a 1, showing the unit is not receiving any 1 pps signals.
>
> What happens is the FC value suddenly jumps from its current setting (-150 for example) to zero, for no apparent reason.
if -150 is really what you saw, then it is out of bounds as the two values returned by fc? are both +ve.
if they go to zero, check the eeprom values
fc!?
58,5485,1534,1434 for example
If they are reasonable then possible the eeprom is not being read correctly after reset.
>
> Then a short while later, the Rubidium's status bytles go crazy, mostly to FF.
This indicates a microprocessor reset . After a power cycle you can see:
st?
255,255,255,243,162,255 ie most FF
when you read again you get the current values
st?
0,0,0,0,130,0 for example
then if you reset the microprocessor
RS 1
PRS_10
st?
255,255,255,243,162,255 back to the initial state
st?
0,0,0,0,130,0
So it looks like you are getting resets. Possibly from brown outs. Maybe a low power condition at reset prevents eprom data being read correctly.
Are the affected RBs on the same power supplies?
>
> Then the rubidium goes back to the previous correct state (all zeros) but the FC value is now zero.
>
> If you do connect a 1 pps, you get the same fault.
>
> The confusing thing is I've suddenly seen this on 3 or 4 different rubidium's. Some brand new, some five years old.
>
> Obviously suspect the power supply, but it doesn't seem to be that.
>
> Best Regards
>
> Martyn
>
>
> _______________________________________________
> time-nuts mailing list -- time-nuts at febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
"The power of accurate observation is commonly called cynicism by those who have not got it. »
George Bernard Shaw
More information about the Time-nuts_lists.febo.com